Assertion (A) : The calorimetric method of pH measurement is based on the assumption that if an indicator has the same colour in two solutions, then the pH of both the solutions are same.

Reason (R) : The colour developed depends not only on the pH but also on other factors.

This question was previously asked in
UGC NET 2023 Electronic Science Question Paper (13-Dec-2023) (Shift 1)
The correct answer is

Both (A) and (R) are correct and (R) is the correct explanation of (A)

 Both statements are true, and the reason explains precisely why the assertion has to be phrased as an assumption rather than as a fact — option 1.

What the colorimetric method does. An indicator is a weak acid whose acid and base forms differ in colour, and the ratio of the two forms is governed by the pH:

\(pH=pK_{a}+\log_{10}\dfrac{\left[\text{In}^{-}\right]}{\left[\text{HIn}\right]}\)

Matching the observed colour against a chart of standards therefore gives an estimate of pH — and the whole procedure rests on taking equal colour to mean equal pH.

Why that is only an assumption. Colour is not determined by pH alone, and several other influences act on it:

FactorEffect on the reading
TemperatureShifts the indicator's pKa
Ionic strength / salt errorAlters activities, so the same pH gives a different ratio
Coloured or turbid samplesMask or distort the indicator's own colour
Oxidising or reducing agentsBleach or chemically alter the dye
Protein errorThe dye binds to protein and changes colour
Indicator concentrationThe dye itself perturbs a weakly buffered sample

The logical relationship is what makes the answer option 1. Notice that the reason does not contradict the assertion — it accounts for its status. Precisely because colour depends on more than pH, the identity of colour cannot be derived; it has to be assumed, and it holds only when those other factors are matched between sample and standard. The reason therefore tells us why the assertion is an assumption at all, which is exactly what "correct explanation" means here.

What follows in practice. The method is quick and cheap and needs no power, which is why indicator papers remain useful for a rough check. But its accuracy is only about ±0.5 pH, and for anything better the potentiometric method is used instead — a glass electrode developing

\(E=E_{0}-\dfrac{2.303RT}{F}\,pH\)

about 59 mV per pH unit at 25 °C. That method depends on a potential rather than a perception, so it is unaffected by colour and turbidity, and its temperature dependence is explicit in the equation and can be compensated.

Hence, both (A) and (R) are correct and (R) is the correct explanation of (A).

Was this answer helpful?

Similar Questions

  1. Match the following :

    List – IList – II  
    a. Miller sweep in CROi. storage oscilloscope
    b. Study of transientsii. integrator
    c. Korotkoff soundiii. phase measurement
    d. Lissajous patterniv. blood pressure

    Codes :

  2. Assertion (A) : Oscilloscope provides graphical representation of time varying signals.

    Reason (R) : Bandwidth is the limitation of oscilloscope.

  3. In a normal ECG wave form, which wave has the maximum amplitude ?

  4. Match the following :

    List – I List – II 
    a. LEDi. Heavily doped
    b. Avalanche Photodiodeii. Coherent radiation
    c. Tunnel diodeiii. Spontaneous emission
    d. LASERiv. Current gain

    Codes :

  5. Match the following :

    List – IList – II  
    a. Storage oscilloscopei. Pulse testing
    b. Sampling oscilloscopeii. Time domain
    c. High frequency oscilloscopeiii. Long term memories
    d. Low frequency oscilloscopeiv. Servo system analyzer

    Codes :

  6. Assertion (A) : The shift in the position of the bright spot on the CRO screen per unit change of voltage across the deflection plates is called deflection sensitivity.

    Reason (R) : The path traveled by the spot on a CRO screen during the retrace time is called retrace path.

  7. A current of 300 µA produces FSD in an instrument whose coil resistance is 75 Ω. What will be the value of shunt resistors if the instrument is to measure current of 5A ?

  8. The resistor value required to use a 0-1 mA meter with an internal resistance of 125 Ω for a 0-1 V meter will be :

  9. There are number of distinct advantages of the digital storage oscilloscope as given below :

    (A) A waveform to be stored is digitized
    (B) The waveform is stored in a digital memory
    (C) The stored waveform can be displayed as and when required
    (D) The cost is reduced as compared to conventional CRT

    Choose the correct answer from the options given below :

  10. Read the following statements with respect to a Wein bridge :

    (A) It is an AC bridge to measure frequency
    (B) It is a DC bridge to measure amplitude
    (C) It is used as frequency determining element
    (D) It is used as band pass filter
    (E) It is used as Notch filter

    Choose the correct answer from the options given below :


Important Questions from Miscellaneous

  1. The magazine in which Mahatma Gandhi mentioned what he wanted the Constitution to do is:

  2. Which gas shields the surface of the earth from ultraviolet radiation from the sun?

  3. Which event is marked as an Intangible Cultural Heritage of Humanity by UNESCO?

  4. Who has been conferred with the rank of the Commander of the Order of the British Empire in 2018?

  5. Who directead the film ‘Bhuvan Shome’?

Need Expert Advice?
Upcoming Exams
MH SET
October 25, 2026
CTET
December 12, 2026
Test Series
UGC NET img
Teaching
UGC NET (Paper 1) 2026 Mock Test Series
476 Tests 1 Tests Free
4.3(72)
English, Hindi

Start Your Preparation with Prepp Mobile App

Download the app from Google Play & App Store
Download the app from Google Play & App Store
Prepp Mobile App