All Exams Test series for 1 year @ ₹349 only
Question

Assertion (A) : Oscilloscope provides graphical representation of time varying signals.

Reason (R) : Bandwidth is the limitation of oscilloscope.

This question was previously asked in
UGC NET 2014 Paper 2 Computer Science and Applications Question Paper (28-Dec-2014)
The correct answer is

Both (A) and (R) are true, but (R) is not the correct explanation of (A).

 Both statements are true, but one describes what an oscilloscope does and the other what limits it — a limitation is not an explanation of a capability. Option 2.

The assertion is the instrument's definition. An oscilloscope plots voltage against time: the vertical amplifier deflects the trace in proportion to the input while the timebase sweeps it horizontally at a known rate, so what appears is the waveform itself — rise time, overshoot, ringing, duty cycle, glitches. That is what distinguishes it from a voltmeter, which collapses the whole waveform into a single number.

The reason is also true. Bandwidth is the headline specification of any scope, defined as the frequency at which a sine wave's displayed amplitude has fallen by 3 dB — to 70.7 % of its true value. Above that the instrument understates what it is shown.

How bandwidth limits what can be trusted. For a scope with a Gaussian response the rise time follows

\(t_{r}\approx\dfrac{0.35}{BW}\)

so a 100 MHz instrument cannot display an edge faster than about 3.5 ns. Worse, the measured rise time is the combination of signal and instrument:

\(t_{measured}=\sqrt{t_{signal}^{2}+t_{scope}^{2}}\)

which is why the usual rule is to choose a scope with three to five times the bandwidth of the fastest edge of interest. A square wave needs several odd harmonics to keep its shape, so a 100 MHz square wave viewed on a 100 MHz scope appears as a sine.

Why the two statements do not connect. The reason would explain the assertion only if bandwidth were the cause of the graphical display. It is not: bandwidth constrains how faithfully the graph is drawn. The instrument would still plot voltage against time if its bandwidth were ten times larger or ten times smaller. Both sentences are correct and both are about oscilloscopes, which is precisely the situation option 2 describes.

Other limitations deserve mention alongside it — sample rate and memory depth on a digital scope, vertical resolution of eight bits on most, probe loading, and the trigger's own bandwidth — none of which explains the assertion either.

Hence, both are true, but (R) is not the correct explanation of (A).

Was this answer helpful?

Similar Questions

  1. Match the following :

    List – IList – II  
    a. Miller sweep in CROi. storage oscilloscope
    b. Study of transientsii. integrator
    c. Korotkoff soundiii. phase measurement
    d. Lissajous patterniv. blood pressure

    Codes :

  2. In a normal ECG wave form, which wave has the maximum amplitude ?

  3. Match the following :

    List – I List – II 
    a. LEDi. Heavily doped
    b. Avalanche Photodiodeii. Coherent radiation
    c. Tunnel diodeiii. Spontaneous emission
    d. LASERiv. Current gain

    Codes :

  4. Match the following :

    List – IList – II  
    a. Storage oscilloscopei. Pulse testing
    b. Sampling oscilloscopeii. Time domain
    c. High frequency oscilloscopeiii. Long term memories
    d. Low frequency oscilloscopeiv. Servo system analyzer

    Codes :

  5. Assertion (A) : The shift in the position of the bright spot on the CRO screen per unit change of voltage across the deflection plates is called deflection sensitivity.

    Reason (R) : The path traveled by the spot on a CRO screen during the retrace time is called retrace path.

  6. A current of 300 µA produces FSD in an instrument whose coil resistance is 75 Ω. What will be the value of shunt resistors if the instrument is to measure current of 5A ?

  7. An ammeter of 1 mA possesses a resistance of 100 ohms. This ammeter is to be converted into 1A ammeter. The value of shunt resistance required will be :

  8. For a Lissajous pattern as shown in Figure below, determine the frequency of vertical signal if the frequency of horizontal signal is 12 kHz.

  9. Write them in descending order of measurement units

    A. atto

    B. tera

    C. Kilo

    D. micro

    E. pico

    Choose the correct answer from the options given below :

  10. Followings is the sequence of blocks in spectrum Analyser :

    A. Detector & video Amplifier

    B. Fixed Frequency Active filter

    C. Attenuator (Adjustable)

    D. Mixer & Tunable Detector

    Choose the correct answer from the options given below :


Important Questions from Miscellaneous

  1. Which of the following is NOT a state function or state variable in a thermodynamic system?

  2. What is the full form for "BPL" in Socio-economic aspects?
  3. In October 2024, which of the following companies was awarded the contract for design, manufacturing and commissioning of two high-speed trainsets by Integral Coach Factory (ICF) to be used in India's first bullet train between Mumbai and Ahmedabad?
  4. Which of the following methods of reproduction simply involves the splitting of the parent cell into two daughter cells in the bacteria?
  5. A, B, C and D opened shopping complexes in the same city. The shopping complex of A is 6 km west of the shopping complex of B. The shopping complex of C is 4 km north of the shopping complex of B. The shopping complex of D is 6 km west of the shopping complex of C. In which direction is the shopping complex of D from the shopping complex of A?
Need Expert Advice?
Upcoming Exams
GSET
November 29, 2026
CTET
December 12, 2026
WB SET
December 13, 2026
CBSE UGC NET
December 14, 2026
CSIR UGC NET
December 19, 2026
Test Series
UGC NET img
Teaching
UGC NET (Paper 2) December 2026 Mock Test Series
799 Tests 2 Tests Free
252 Attempts
4.4(43)
English, Hindi
More Questions from UGC NET

Start Your Preparation with Prepp Mobile App

Download the app from Google Play & App Store
Download the app from Google Play & App Store
Prepp Mobile App