Match the following :

List – IList – II  
a. Storage oscilloscopei. Pulse testing
b. Sampling oscilloscopeii. Time domain
c. High frequency oscilloscopeiii. Long term memories
d. Low frequency oscilloscopeiv. Servo system analyzer

Codes :

This question was previously asked in
UGC NET 2014 Paper 1 Question Paper (28-Dec-2014)
The correct answer is

a-iii, b-ii, c-i, d-iv

 The storage oscilloscope's pairing is beyond doubt and selects the code: a-iii, b-ii, c-i, d-iv — option 2.

InstrumentMatch
a. Storage oscilloscopeiii. Long term memories
b. Sampling oscilloscopeii. Time domain
c. High frequency oscilloscopei. Pulse testing
d. Low frequency oscilloscopeiv. Servo system analyzer

a — why storage means memory. An ordinary oscilloscope's trace is redrawn many times a second and persists only as long as the phosphor glows. A single, non-repetitive event flashes once and is gone. A storage oscilloscope retains it — originally on a bistable storage mesh inside the CRT, now by digitising into memory — and holds it indefinitely for examination. "Long term memories" is a direct description of that capability, and no other instrument in the list has it.

c — why pulse testing needs bandwidth. A pulse's edges contain its high-frequency content: for a Gaussian response,

\(t_{r}\approx\dfrac{0.35}{BW}\)

so an instrument of 100 MHz cannot display an edge faster than 3.5 ns, and the measured rise time combines signal and instrument in quadrature. Examining overshoot, ringing and settling on fast pulses is therefore squarely the high-frequency oscilloscope's work.

d — why servo work is low-frequency. A servo's mechanical loop responds over hundreds of milliseconds to seconds, so its step response, overshoot and settling time occur at fractions of a hertz to a few hertz. Displaying that needs a very slow timebase and DC coupling — the opposite requirement from pulse work.

b — the sampling oscilloscope. It builds a picture of a repetitive waveform by taking one sample per cycle, each slightly later than the last, and assembling them into a complete trace — equivalent-time sampling. This reaches far beyond the bandwidth of any real-time instrument, into tens of gigahertz, because only the sampling gate need be fast. It reconstructs the waveform against time, so "time domain" fits, though the phrase is the loosest in the list — which is why the answer is flagged.

The pairing is nevertheless secure, since a is certain and the remaining three sort themselves by frequency range.

Hence, the correct code is a-iii, b-ii, c-i, d-iv.

Was this answer helpful?

Similar Questions

  1. Match the following :

    List – IList – II  
    a. Miller sweep in CROi. storage oscilloscope
    b. Study of transientsii. integrator
    c. Korotkoff soundiii. phase measurement
    d. Lissajous patterniv. blood pressure

    Codes :

  2. Assertion (A) : Oscilloscope provides graphical representation of time varying signals.

    Reason (R) : Bandwidth is the limitation of oscilloscope.

  3. In a normal ECG wave form, which wave has the maximum amplitude ?

  4. Match the following :

    List – I List – II 
    a. LEDi. Heavily doped
    b. Avalanche Photodiodeii. Coherent radiation
    c. Tunnel diodeiii. Spontaneous emission
    d. LASERiv. Current gain

    Codes :

  5. Assertion (A) : The shift in the position of the bright spot on the CRO screen per unit change of voltage across the deflection plates is called deflection sensitivity.

    Reason (R) : The path traveled by the spot on a CRO screen during the retrace time is called retrace path.

  6. A current of 300 µA produces FSD in an instrument whose coil resistance is 75 Ω. What will be the value of shunt resistors if the instrument is to measure current of 5A ?

  7. An ammeter of 1 mA possesses a resistance of 100 ohms. This ammeter is to be converted into 1A ammeter. The value of shunt resistance required will be :

  8. For a Lissajous pattern as shown in Figure below, determine the frequency of vertical signal if the frequency of horizontal signal is 12 kHz.

  9. Write them in descending order of measurement units

    A. atto

    B. tera

    C. Kilo

    D. micro

    E. pico

    Choose the correct answer from the options given below :

  10. Followings is the sequence of blocks in spectrum Analyser :

    A. Detector & video Amplifier

    B. Fixed Frequency Active filter

    C. Attenuator (Adjustable)

    D. Mixer & Tunable Detector

    Choose the correct answer from the options given below :


Important Questions from Miscellaneous

  1. Which event is marked as an Intangible Cultural Heritage of Humanity by UNESCO?

  2. Who has been conferred with the rank of the Commander of the Order of the British Empire in 2018?

  3. Who directead the film ‘Bhuvan Shome’?

  4. The book “In my mother’s house” is written by _______

  5. In which state is Vishwa Mitra Ka Tila situated?

Need Expert Advice?
Upcoming Exams
CTET
December 12, 2026
Test Series
UGC NET img
Teaching
UGC NET (Paper 2) 2026 Mock Test Series
796 Tests 2 Tests Free
4.4(39)
English
More Questions from UGC NET

Start Your Preparation with Prepp Mobile App

Download the app from Google Play & App Store
Download the app from Google Play & App Store
Prepp Mobile App