A metallic strain-gauge (SG) with resistance $R_{SG}$ is connected as shown in the figure, where $R_{L1}$, $R_{L2}$, $R_{L3}$ represent the lead wire resistances. The SG has a gauge factor of 2 and nominal resistance $R_N$ of 125 $\Omega$. When the SG is subjected to a tensile strain of $2 \times 10^{-3}$, the resulting change in $R_{SG}$ is $\Delta R$. The $\Delta R$ value is measured as $\Delta R_{MEAS} = R_{EQ2} - R_{EQ1}$. The $R_{EQ1}$ and $R_{EQ2}$ are the equivalent resistances measured between the terminals 1 and 2, and terminals 2 and 3, respectively.
If $R_{L1} = R_{L2} = 5 \ \Omega$, and $R_{L3} = 4.95 \ \Omega$, the measured value of tensile strain is ______ $\times 10^{-3}$ (rounded off to two decimal places).
The problem involves calculating the measured tensile strain using a strain-gauge circuit. The given parameters are:
The change in strain-gauge resistance is given by:
\[\Delta R = GF \times R_N \times \varepsilon\]
Substituting the values:
\[\Delta R = 2 \times 125 \times 2 \times 10^{-3} = 0.5 \, \Omega\]
The measured resistances are:
\( R_{EQ1} = R_N + R_{L1} + R_{L2} = 125 + 5 + 5 = 135 \, \Omega \)
\( R_{EQ2} = R_{SG} + R_{L2} + R_{L3} = (R_N + \Delta R) + 5 + 4.95 \)
\( R_{EQ2} = (125 + 0.5) + 5 + 4.95 = 135.45 \, \Omega \)
The measured change in resistance:
\(\Delta R_{MEAS} = R_{EQ2} - R_{EQ1} = 135.45 - 135 = 0.45 \, \Omega\)
The measured tensile strain, \( \varepsilon_{MEAS} \), is calculated using:
\(\varepsilon_{MEAS} = \frac{\Delta R_{MEAS}}{GF \times R_N}\)
Substitute the values:
\(\varepsilon_{MEAS} = \frac{0.45}{2 \times 125} = 0.0018 = 1.8 \times 10^{-3}\)
The calculated strain is 1.8 which falls within the given range of 1.75 to 1.85.
Which of the following is NOT an advantage of LVDT?
Flow can be measured by ____.
In the force transducer shown in Figure (a), four identical strain gauges S1, S2, S3, and S4 are mounted on a cantilever at equal distance from its base. S1 and S2 are mounted on the top surface and S3 and S4 are mounted on the bottom surface, as shown in the Figure (a). These strain gauges are to be connected to form a Wheatstone bridge consisting of four arms A, B, C, and D, as shown in the Figure (b). From the following options, the correct order to maximize the measurement sensitivity is