This solution details the calculation for determining the holding time required for sterilizing an industrial fermentor, based on spore death kinetics.
First, calculate the total number of spores initially present in the fermentor.
The number of surviving spores after time $t$ is governed by first-order kinetics: $N(t) = N_0 e^{-kt}$. The Sterility Assurance Level (SAL) aims to minimize the probability of contamination. For a large initial population, the probability of contamination ($P_{contam}$) is approximately equal to the expected number of survivors ($E[N_{survivors}]$).
The sterilization requirement is that the expected number of survivors must be less than or equal to the target contamination probability:
$E[N_{survivors}] \le P_{contam}$Substituting the expression for $E[N_{survivors}] = N_{0, \text{total}} e^{-kt}$:
$N_{0, \text{total}} e^{-kt} \le P_{contam}$ $10^{13} \text{ spores} \times e^{-kt} \le 10^{-3}$This simplifies to finding the required reduction factor:
$e^{-kt} \le \frac{10^{-3}}{10^{13}} = 10^{-16}$Solve for the holding time ($t$) using the death rate constant ($k$).
The calculated minimum holding time is approximately $9.2104$ minutes. The question requires this value to be rounded off to the nearest integer.
Rounding $9.2104$ to the nearest integer yields $9$ minutes.
Therefore, the holding time of the sterilization process is 9 min (rounded off to the nearest integer).
The decimal reduction time of a microbe during sterilization at $120 \text{ °C}$ with a first order thermal death rate constant of $1 \text{ min}^{-1}$ will be _______________ min (rounded off to 1 decimal place).
Moist heat sterilization of spores at $121 \text{ } ^\circ C$ follows first order kinetics as per the expression:
$ \frac{dN}{dt} = -k_d N $
where, N is the number of viable spores, t is the time, $k_d$ is the rate constant and $ \frac{dN}{dt} $ is the rate of change of viable spores.
If $k_d$ value is $1.0 \text{ min}^{-1}$, the time (in minutes) required to reduce the number of viable spores from an initial value of $10^{10}$ to a final value of 1 is (up to two decimal places)______.
Decimal reduction time of a bacterial strain is $20$ min. Specific death rate constant in $min^{-1}$ (rounded off to two decimal places) is____.