All Exams Test series for 1 year @ ₹349 only
Question

An ammeter of 0-50 A range has an accuracy of ± 1% of full scale reading. The current measured is 10 A. The uncertainty in measured value is

The correct answer is

± 5%

Ammeter Uncertainty Calculation Explained

This solution explains how to calculate the uncertainty in a current measurement using an ammeter, considering its accuracy specifications.

Understanding the Problem

We are given an ammeter with the following details:

  • Range: 0-50 A (This means the Full Scale Reading, FSR, is 50 A)
  • Accuracy: $\pm 1\%$ of Full Scale Reading (FSR)
  • Measured Current: 10 A

The goal is to find the uncertainty in the measured value of 10 A, expressed as a percentage of the measured value itself.

Calculating Absolute Uncertainty

The accuracy specification tells us the maximum error relative to the full scale reading. First, we calculate the absolute uncertainty:

Absolute Uncertainty = Accuracy Percentage $\times$ Full Scale Reading (FSR)

Absolute Uncertainty = $1\% \times 50 \, \text{A}$

Absolute Uncertainty = $\frac{1}{100} \times 50 \, \text{A}$

Absolute Uncertainty = $0.5 \, \text{A}$

This means the ammeter reading can be off by as much as $0.5 \, \text{A}$ from the true value.

Determining Percentage Uncertainty of Measured Value

The question asks for the uncertainty relative to the *measured value* (10 A), not the full scale reading.

Percentage Uncertainty = $ \left( \frac{\text{Absolute Uncertainty}}{\text{Measured Value}} \right) \times 100\% $

We substitute the values we have:

Percentage Uncertainty = $ \left( \frac{0.5 \, \text{A}}{10 \, \text{A}} \right) \times 100\% $

Percentage Uncertainty = $ 0.05 \times 100\% $

Percentage Uncertainty = $ 5\% $

Final Uncertainty

Therefore, the uncertainty in the measured value of 10 A is $\pm 5\%$.

Summary Table

Parameter Value Calculation/Notes
Ammeter Range (FSR) 50 A Given
Accuracy Specification $\pm 1\%$ of FSR Given
Absolute Uncertainty $0.5 \, \text{A}$ $1\% \times 50 \, \text{A}$
Measured Current 10 A Given
Uncertainty (% of Measured Value) $\pm 5\%$ $(0.5 \, \text{A} / 10 \, \text{A}) \times 100\%$

Was this answer helpful?

Important Questions from Error Analysis

  1. Errors that occur after taking care of all gross and systematic errors are called as:
  2. A 0-200 V voltmeter has an accuracy of 0.75% of full-scale reading. If the voltage measured is 100 V, the error is

  3. Suppose that resistors R1 and R2 are connected in parallel to give an equivalent resistor R. If resistors R1 and R2 have tolerance of 1% each the equivalent resistor R for resistor R1 = 300 Ω and R2 = 200 Ω will have tolerance of

  4. A \(4 \frac 1 2\) digit DMM has the error specification as 0.2% of reading +10 counts. If a d.c. voltage of 100 V is read on its 200 V full-scale, the maximum error that can be expected in the reading is

  5. A voltage V1 is measured 100 times and its average and standard deviation are 100 V and 1.5 V respectively. A second voltage V2, which is independent of V1, is measured 200 times and its average and standard deviation are 150 V and 2 V respectively. V3 is computed as: V3 + V1 + V2. Then the standard deviation of V3 in volts is ________

Need Expert Advice?

Start Your Preparation with Prepp Mobile App

Download the app from Google Play & App Store
Download the app from Google Play & App Store
Prepp Mobile App