Two ammeters x and y have resistances of 1.2 Ω and 1.5 Ω respectively and they give full scale deflection with 150 mA and 250 mA respectively. The ranges have been extended by connecting shunts so as to give full scale deflection with 15 A. The ammeters along with shunts are connected in parallel and then placed in a circuit in which the total current flowing is 15 A. The current in amperes indicated in ammeter x is-
Concept:
The shunts and multipliers are used in the extension of the range of voltmeters and ammeters.
For ammeter, a low shunt resistance (Rsh ≪≪ Rm) is connected to increase its range
\({{\bf{R}}_{{\bf{sh}}}} = \frac{{{{\bf{R}}_{\bf{m}}}}}{{{\bf{m}} - 1}}\)
Rsh = shunt resistance
Rm = meter resistance
m = I / Im = multiplying factor or multiplying power
For the voltmeter, a high series resistance (Rse ≫≫Rm) is connected to increase its range
\({{\bf{R}}_{{\bf{se}}}} = {{\bf{R}}_{\bf{m}}}\left( {{\bf{m}} - 1} \right)\)
\({\rm{m}} = \frac{{\rm{V}}}{{{{\rm{V}}_{\rm{m}}}}}\) = multiplying power/factor
Calculation:

Given meter resistance
Rmx = 1.2 Ω, Rmy = 1.5 Ω
Shunt resistances need to be connected to increase the range of the meters to 15 A
Imx = 150 mA, Imy = 250 mA
\(\begin{array}{l} {R_{shx}} = \frac{{1.2}}{{\left( {\frac{{15}}{{150 \times {{10}^{ - 3}}}} - 1} \right)}} = 0.0121\;{\rm{Ω }}\\ {R_{shy}} = \frac{{1.5}}{{\left( {\frac{{15}}{{250 \times {{10}^{ - 3}}}} - 1} \right)}} = 0.02542\;{\rm{Ω }} \end{array}\)
Equivalent resistance offered by meter x
Rx = Rmx || Rshx = 0.012 Ω
Equivalent resistance offered by meter y
Ry = Rmy || Rshy = 0.025 Ω
Current through ammeter ‘x’ is
\(I = \left( {\frac{{0.025\times15}}{{0.012 + 0.025}}} \right) = 10.13\; A\)