The true r.m.s. voltmeter employes two thermocouples in order to -
cancel out the nonlinear effects of first thermocouple
A true RMS voltmeter is an instrument designed to measure the true Root Mean Square (RMS) value of an AC voltage, regardless of its waveform. Unlike average-responding or peak-responding voltmeters, which are calibrated for sinusoidal waveforms and provide inaccurate readings for non-sinusoidal inputs, a true RMS voltmeter provides an accurate RMS value for complex waveforms such as square waves, triangular waves, pulse trains, and distorted sinusoids.
The RMS value of a voltage or current waveform is significant because it represents the heating effect or power dissipation capability of the waveform in a resistive load. For a voltage \(V(t)\), the RMS value \(V_{\text{rms}}\) is defined as:
$$ V_{\text{rms}} = \sqrt{\frac{1}{T} \int_{0}^{T} V^2(t) \, dt} $$
This definition shows that the RMS value is essentially related to the average power of the signal, which is proportional to the square of the voltage.
Many true RMS voltmeters, especially older or high-precision analog ones, employ thermal converters, which are typically based on thermocouples. A thermocouple generates a voltage proportional to a temperature difference. In a thermal converter, the AC voltage to be measured is applied across a heating element (often a resistor). The heat generated by this element is proportional to the square of the input voltage (\(P \propto V^2\)). This heat then raises the temperature of a thermocouple junction, which produces a DC voltage output.
The main challenge with a single thermocouple is that its output voltage is not perfectly linear with respect to temperature, and its response can be affected by ambient temperature changes (drift). This non-linear characteristic and susceptibility to drift can significantly affect the accuracy of the RMS measurement.
To overcome the inherent non-linear effects of a single thermocouple and to improve accuracy, true RMS voltmeters often utilize two thermocouples in a carefully designed circuit, typically a self-balancing potentiometer or a differential arrangement. This dual-thermocouple setup is crucial for high-precision true RMS measurements.
Here's how the two thermocouples work:
The system operates as a feedback loop:
By operating both thermocouples at the same temperature, the non-linear effects of the first thermocouple are effectively cancelled out by the equally non-linear but opposing response of the second thermocouple. Any non-linearity in the thermocouple's voltage-to-temperature characteristic affects both thermocouples similarly, and thus cancels out in the differential measurement. This also effectively compensates for ambient temperature variations (drift) because both thermocouples are exposed to the same ambient conditions.
The voltage output of a thermocouple is not perfectly linear with the temperature difference across its junctions. While it's approximately linear over a small range, significant deviations occur over wider temperature swings. Since the heating element's temperature rise is proportional to the square of the input voltage, any non-linearity in the thermocouple's voltage output would lead to an inaccurate RMS reading for various input levels. The use of a second, identical thermocouple in a balancing circuit ensures that these non-linearities are inherently compensated for, leading to a much more accurate true RMS measurement.
The employment of two thermocouples in a true RMS voltmeter is a sophisticated design choice primarily intended to cancel out the nonlinear effects of the primary measuring thermocouple. This differential arrangement significantly enhances the accuracy of the instrument, making it capable of providing precise RMS measurements even for complex, non-sinusoidal waveforms. While it also helps in preventing drift due to ambient temperature changes, the core reason for using two thermocouples in a feedback loop is to achieve a highly linear and accurate RMS response by nullifying the inherent non-linearity of the thermal sensing elements.