The activity of lactate dehydrogenase can be measured by monitoring the following reaction: 

Pyruvate + NADH $ \longrightarrow $ Lactate + $NAD^+$ 

The molar extinction coefficient of NADH at 340 nm is $6220 \ M^{-1}.cm^{-1}$. $NAD^+$ does not absorb at this wavelength. In an assay, $25 \ \mu L$ of a sample of enzyme (containing $5 \ \mu g$ protein per mL) was added to a mixture of pyruvate and NADH to give a total volume of 3 mL in a cuvette of 1 cm pathlength. The rate of decrease in absorbance at 340 nm was $0.14 \ min^{-1}$. The specific activity of the enzyme will be ____________________ $ \mu mol.min^{-1}.mg^{-1}$.

To calculate the specific activity of lactate dehydrogenase, we need to first determine the reaction rate in terms of $ \mu mol.min^{-1} $ and then normalize it to the enzyme concentration in terms of mg of protein.

Step 1: Calculate the reaction rate in $ \mu mol.min^{-1} $.

Given the decrease in absorbance per minute is $0.14 \ min^{-1}$, we use Beer-Lambert Law:

$ A = \varepsilon \cdot c \cdot l $

Where $A$ is absorbance, $\varepsilon$ is molar extinction coefficient, $c$ is concentration, and $l$ is path length. Given $\varepsilon = 6220 \ M^{-1}.cm^{-1}$ and $l = 1 \ cm$, we find the change in concentration per minute (by setting $c$ equal to the change in concentration, which is $-\Delta [NADH]$ as absorbance decreases):

$ \Delta A = \varepsilon \cdot (-\Delta [NADH]) \cdot 1 $

Thus, $-\Delta [NADH] = \frac{0.14}{6220} = 2.25 \times 10^{-5} \ M.min^{-1}$

Convert this to $ \mu mol.L^{-1}.min^{-1} $:

$2.25 \times 10^{-5} \ mol.L^{-1}.min^{-1} = 22.5 \ \mu mol.L^{-1}.min^{-1} $

Since the total volume is $3 \ mL = 0.003 \ L$, the total reaction rate is:

$ 22.5 \ \mu mol.L^{-1}.min^{-1} \times 0.003 \ L = 0.0675 \ \mu mol.min^{-1} $

Step 2: Calculate the specific activity.

The enzyme solution added was $25 \ \mu L$ of $1 \ mL = 0.025 \ mL$, and the concentration of protein in that solution is $5 \ \mu g.mL^{-1}$:

$ \text{Total protein} = 5 \ \mu g.mL^{-1} \times 0.025 \ mL = 0.125 \ \mu g $

Convert $0.125 \ \mu g$ to mg:

$0.125 \ \mu g = 0.000125 \ mg $

The specific activity is given by the reaction rate divided by the enzyme mass:

$ \text{Specific Activity} = \frac{0.0675 \ \mu mol.min^{-1}}{0.000125 \ mg} = 540 \ \mu mol.min^{-1}.mg^{-1} $

This value falls within the expected range of 525,555, confirming its accuracy.

Conclusion: The specific activity of the enzyme is $540 \ \mu mol.min^{-1}.mg^{-1}$.

Was this answer helpful?

Important Questions from Enzyme Kinetics and Michaelis Menten Equation

  1. The catalytic efficiency of an enzyme following Michaelis-Menten kinetics is defined by
  2. You are characterizing a new enzyme isolated and purified in the laboratory. If the maximum velocity of the enzyme is $1800 \text{ } \mu moles \text{ L}^{-1}  \text{min}^{-1}$ and the total concentration of the enzyme in the reaction mixture is $1.5 \mu \text{M}$, then the turnover number of the enzyme is _______ $\text{s}^{-1}$. (answer in integer)

  3. You have purified an enzyme using a series of chromatographic methods. It was observed that a $10 \mu \text{  g mL}^{-1}$ of this purified enzyme converted $10 \text{ mM}$ substrate per hour at 25$^{\circ}$C and pH 7. Its specific activity is _______ $\text{IU  } \mu\text{g}^{-1}$. (rounded off to three decimal places)

  4. Within the Michaelis-Menten framework, the ratio of $v_0/V_{max}$ 

    when $[S] = 20 \times K_m$ is _________. 

    (Round off to two decimal places)

  5. An enzyme (E) catalyzes the biochemical reaction $A \rightarrow B$ with $k_{cat}$ equal to $500 s^{-1}$. If the initial reaction velocity ($V_0$) is $10 \mu M.s^{-1}$ at the total enzyme concentration $[E_t]$ of 30 nM and substrate concentration $[A]$ of $40 \mu M$, the value of $K_m$ (in $\mu M$) is ________
Need Expert Advice?

Start Your Preparation with Prepp Mobile App

Download the app from Google Play & App Store
Download the app from Google Play & App Store
Prepp Mobile App