Question

During the measurement of voltage, the voltmeter responded with a 0.18-V change when the input was varied by 0.2 V. Find the sensitivity of the instrument.

The correct answer is

0.9

Calculating Voltmeter Sensitivity

The question asks us to determine the sensitivity of a voltmeter based on how much its reading changes in response to a change in the input voltage it is measuring.

Sensitivity of an instrument is a measure of how much the output changes for a given change in the input. It is typically defined as the ratio of the change in output quantity to the change in input quantity that causes it.

Understanding Instrument Sensitivity

For a measuring instrument like a voltmeter, the sensitivity tells us how finely it can register changes in the voltage being measured. A higher sensitivity means a smaller change in input voltage will result in a noticeable change in the voltmeter's reading (output).

Formula for Sensitivity

The general formula for sensitivity (S) is given by:

S = \frac{\text{Change in Output}}{\text{Change in Input}}

In this specific case, for the voltmeter:

\text{Sensitivity} = \frac{\text{Change in Voltmeter Reading}}{\text{Change in Input Voltage}}

Applying the Given Values

From the question, we are given the following information:

  • Change in voltmeter reading (Output change) = 0.18 V
  • Change in input voltage (Input change) = 0.2 V

Now, we can substitute these values into the sensitivity formula:

\text{Sensitivity} = \frac{0.18\text{ V}}{0.2\text{ V}}

Step-by-Step Calculation of Sensitivity

Let's perform the calculation:

\text{Sensitivity} = \frac{0.18}{0.2}

To make the division easier, we can multiply the numerator and denominator by 100 to remove the decimal points:

\text{Sensitivity} = \frac{0.18 \times 100}{0.2 \times 100} = \frac{18}{20}

Now, simplify the fraction:

\text{Sensitivity} = \frac{18 \div 2}{20 \div 2} = \frac{9}{10}

Convert the fraction to a decimal:

\text{Sensitivity} = 0.9

The units cancel out (V/V), so sensitivity is a dimensionless quantity in this context, or it can be considered in terms of V/V, but typically stated as a numerical value.

Result

The calculated sensitivity of the voltmeter is 0.9.

Revision Table: Instrument Sensitivity Calculation

Quantity Value
Change in Output (Voltmeter Reading) 0.18 V
Change in Input (Input Voltage) 0.2 V
Formula for Sensitivity \frac{\text{Change in Output}}{\text{Change in Input}}
Calculated Sensitivity 0.9

Additional Information: Voltmeter Characteristics

Besides sensitivity, other important characteristics of a voltmeter include:

  • Accuracy: How close the measured value is to the true value. Often expressed as a percentage of full scale or reading.
  • Precision: The degree of reproducibility of measurements. How close repeated measurements are to each other.
  • Resolution: The smallest change in the input quantity that the instrument can detect.
  • Range: The set of values of the input quantity that the instrument is designed to measure.
  • Input Impedance: Ideally, a voltmeter should have very high input impedance so that it doesn't draw significant current from the circuit being measured, which could alter the voltage itself.

Sensitivity is a key performance indicator for any measuring instrument as it relates directly to how well it can respond to variations in the quantity being measured.

Was this answer helpful?

Important Questions from Sensitivity

  1. An ammeter requires a change of 3 A in its coil to produce a change in deflection of the pointer by 12 mm. Its sensitivity is

  2. If the value of (1 + GH) is less than 1, then sensitivity is/has _____.
  3. There are 2 systems:

    a. An automatic washing machine

    b. An automatic intensity adjustable light bulb

    Which of these systems will be more sensitive to the variation in system's gain?

  4. Study the given table for resistance values of a platinum thermometer measured at a range of temperatures. Calculate the sensitivity of the measurement of the instrument.

    Resistance (Ω) Temperature (°C)
    200100
    205150
    210200
    215250
  5. If meter A requires 100 mA to give full-scale deflection, meter B requires 50 mA to give scale deflection and meter C requires 80 mA to give full-scale deflection, then the

Need Expert Advice?

Start Your Preparation with Prepp Mobile App

Download the app from Google Play & App Store
Download the app from Google Play & App Store
Prepp Mobile App