All Exams Test series for 1 year @ ₹349 only
Question

If meter A requires 100 mA to give full-scale deflection, meter B requires 50 mA to give scale deflection and meter C requires 80 mA to give full-scale deflection, then the

The correct answer is

Meter B is more sensitive

Meter Sensitivity Explained

Understanding the sensitivity of an electrical meter is crucial in electrical measurements. A meter's sensitivity indicates how much current is required to produce a full-scale deflection (FSD). Essentially, the less current a meter needs to achieve full-scale deflection, the more sensitive it is. A highly sensitive meter can detect smaller changes in the measured quantity.

Sensitivity Definition and Formula

The sensitivity of a meter is generally defined as the reciprocal of the current required for full-scale deflection. While often expressed in ohms per volt ($\Omega/\text{V}$) for voltmeters, in the context of comparing ammeters or galvanometers based on their full-scale deflection current, a meter is considered more sensitive if it requires a smaller current for full-scale deflection.

Mathematically, if sensitivity is inversely proportional to the full-scale deflection current ($\text{I}_{\text{FSD}}$), we can express it as:

$\text{Sensitivity} \propto \frac{1}{\text{I}_{\text{FSD}}}$

This relationship implies that a lower full-scale deflection current ($\text{I}_{\text{FSD}}$) corresponds to higher meter sensitivity.

Analyzing Meter Deflection Currents

Let's examine the full-scale deflection (FSD) current requirements for each given meter:

  • Meter A requires $\text{100 mA}$ to give full-scale deflection.
  • Meter B requires $\text{50 mA}$ to give full-scale deflection.
  • Meter C requires $\text{80 mA}$ to give full-scale deflection.

Comparing Meter Sensitivity

To determine which meter is more sensitive, we must compare the full-scale deflection currents for each meter. As established, the meter that requires the least amount of current to achieve full-scale deflection is the most sensitive meter.

Meter Current for Full-Scale Deflection ($\text{I}_{\text{FSD}}$)
Meter A $\text{100 mA}$
Meter B $\text{50 mA}$
Meter C $\text{80 mA}$

From the comparison of the currents required for full-scale deflection:

  • Meter A needs the highest current of $\text{100 mA}$.
  • Meter B needs the lowest current of $\text{50 mA}$.
  • Meter C needs an intermediate current of $\text{80 mA}$.

Since $\text{50 mA}$ is the smallest current required for full-scale deflection among the three meters, Meter B exhibits the highest sensitivity.

Conclusion on Meter Sensitivity

Based on the analysis of the full-scale deflection current for each meter, Meter B requires the least current ($\text{50 mA}$) to achieve full-scale deflection. This makes Meter B the most sensitive among Meter A ($\text{100 mA}$) and Meter C ($\text{80 mA}$). Therefore, Meter B is more sensitive.

Was this answer helpful?

Important Questions from Sensitivity

  1. An ammeter requires a change of 3 A in its coil to produce a change in deflection of the pointer by 12 mm. Its sensitivity is

  2. During the measurement of voltage, the voltmeter responded with a 0.18-V change when the input was varied by 0.2 V. Find the sensitivity of the instrument.

  3. If the value of (1 + GH) is less than 1, then sensitivity is/has _____.
  4. There are 2 systems:

    a. An automatic washing machine

    b. An automatic intensity adjustable light bulb

    Which of these systems will be more sensitive to the variation in system's gain?

  5. Study the given table for resistance values of a platinum thermometer measured at a range of temperatures. Calculate the sensitivity of the measurement of the instrument.

    Resistance (Ω) Temperature (°C)
    200100
    205150
    210200
    215250
Need Expert Advice?

Start Your Preparation with Prepp Mobile App

Download the app from Google Play & App Store
Download the app from Google Play & App Store
Prepp Mobile App