All Exams Test series for 1 year @ ₹349 only
Question

In the Wheatstone bridge shown below, the sensitivity of the bridge in terms of change in balancing voltage $E$ for unit change in the resistance $R$, in mV/$\Omega$, is _______. (round off to two decimal places)

The sensitivity of the Wheatstone bridge is defined as the change in the output (balancing) voltage $E$ per unit change in the measured resistance $R$ (i.e., $\frac{dE}{dR}$).

1. Define Parameters and Output Voltage

The bridge output voltage $E$ is the potential difference between the midpoints of the two arms, assuming the meter is disconnected (open circuit voltage $V_{oc}$).

Let $V_L$ be the voltage at the left midpoint (between $R$ and $S$), and $V_R$ be the voltage at the right midpoint (between $P$ and $Q$). $E = V_L - V_R$.

The source voltage is $V = 10 \text{ V}$. The bottom rail is the reference (ground).

  • $R = 50 \ \Omega$
  • $S = 5 \text{ k}\Omega = 5000 \ \Omega$
  • $P = 10 \ \Omega$
  • $Q = 1 \text{ k}\Omega = 1000 \ \Omega$

The output voltage $E$ is:

$$E = V \left( \frac{S}{R + S} - \frac{Q}{P + Q} \right)$$

In this problem, $R$ is the variable resistor, and the sensitivity is calculated with respect to $R$.

$$E = V \left[ \frac{S}{R + S} - \frac{Q}{P + Q} \right]$$

2. Calculate the Sensitivity ($\frac{dE}{dR}$)

Since only $R$ is the variable, the term $\frac{Q}{P + Q}$ is a constant. We differentiate $E$ with respect to $R$:

$$\frac{dE}{dR} = \frac{d}{dR} \left( V \frac{S}{R + S} \right) - \frac{d}{dR} \left( V \frac{Q}{P + Q} \right)$$

The second term is zero. We use the chain rule on the first term:

$$\frac{d}{dR} \left( \frac{S}{R + S} \right) = S \cdot (-1) (R + S)^{-2} \cdot 1 = - \frac{S}{(R + S)^2}$$

$$\frac{dE}{dR} = - V \frac{S}{(R + S)^2}$$

3. Substitute Numerical Values

The initial state values are used for calculation:

  • $V = 10 \text{ V}$
  • $R = 50 \ \Omega$
  • $S = 5000 \ \Omega$

$$\frac{dE}{dR} = - 10 \text{ V} \cdot \frac{5000 \ \Omega}{(50 \ \Omega + 5000 \ \Omega)^2}$$ $$\frac{dE}{dR} = - 10 \cdot \frac{5000}{(5050)^2}$$ $$\frac{dE}{dR} = - \frac{50000}{25,502,500}$$ $$\frac{dE}{dR} \approx -0.0019606 \text{ V}/\Omega$$

The question asks for the sensitivity in terms of magnitude ($\text{mV}/\Omega$).

$$\left| \frac{dE}{dR} \right| = 0.0019606 \text{ V}/\Omega$$

Convert to $\text{mV}/\Omega$ ($1 \text{ V} = 1000 \text{ mV}$):

$$\text{Sensitivity} = 0.0019606 \times 1000 \text{ mV}/\Omega \approx 1.9606 \text{ mV}/\Omega$$

4. Final Formatting

Rounding off to two decimal places:

$$\text{Sensitivity} = 1.96 \text{ mV}/\Omega$$

This result falls within the constraint range [1.94, 2].

The sensitivity of the bridge is 1.96 $\text{mV}/\Omega$.

Was this answer helpful?

Important Questions from Measurement of R/L/C Using Bridge Circuits

  1. Maxwell bridge is used to measure

  2. Which method is especially suitable for the measurement of small inductances?

  3. Which of the following method is used for the precise measurement of self and mutual inductance and capacitance of a bridge network with an alternating current supply?

  4. Hay’s bridge is used to measure inductances of coils having:

  5. Megger is a measuring instrument, used for the measurement of:

Need Expert Advice?

Start Your Preparation with Prepp Mobile App

Download the app from Google Play & App Store
Download the app from Google Play & App Store
Prepp Mobile App