Question

______ eliminates errors due to contacts and ______ in bridge measuring instruments.

The correct answer is

Kelvin's Double bridge, lead resistances

Kelvin's Double Bridge: Addressing Measurement Errors

Bridge measuring instruments are fundamental tools in electrical engineering used for precise measurement of resistance, inductance, and capacitance. However, when measuring very low resistances (typically below 1 Ohm), certain challenges arise, primarily due to the resistance of the connecting wires (leads) and the contact points themselves. These resistances can introduce significant errors, making accurate measurement difficult.

Understanding Kelvin's Double Bridge

The Kelvin's Double Bridge is a specialized type of bridge circuit specifically designed for the accurate measurement of very low resistances. It is an adaptation of the Wheatstone bridge, but with crucial modifications that enable it to overcome the limitations faced by a standard Wheatstone bridge when dealing with resistances close to or less than 1 Ohm.

Common Errors in Low Resistance Measurement

When attempting to measure very low resistances using conventional bridge circuits like the Wheatstone bridge, two primary sources of error become prominent:

  • Contact Errors: These errors arise from the resistance at the points where the measuring leads make contact with the unknown resistor and the standard resistor. Even good contacts can have a resistance of several micro-ohms, which becomes significant when measuring milliohm or micro-ohm resistances.
  • Lead Resistances: These refer to the inherent resistance of the connecting wires (leads) themselves. Although the resistance of short, thick wires might seem negligible, when the unknown resistance is very small, the lead resistance can be a substantial fraction of, or even greater than, the resistance being measured. This resistance in the leads connected to the galvanometer can introduce inaccuracies.

In low resistance measurements, these unwanted resistances (contact and lead resistances) can distort the measurement, leading to inaccurate results because the voltage drop across them becomes comparable to the voltage drop across the actual low resistance being measured.

Eliminating Errors with Kelvin's Double Bridge

The Kelvin's Double Bridge is ingeniously designed to eliminate errors caused by both contact resistances and lead resistances. It achieves this through a unique four-terminal connection for the unknown and standard resistors and the inclusion of additional ratio arms:

  • Four-Terminal Resistors: Unlike two-terminal resistors, which bundle current and potential connections, four-terminal resistors have separate pairs of terminals for current injection and voltage sensing. The Kelvin's Double Bridge utilizes these, ensuring that the current flows through one set of terminals, while the voltage drop across the resistor itself is measured across a separate pair of potential terminals. This separation prevents the contact resistance at the current terminals from affecting the voltage measurement.
  • Double Ratio Arms: The bridge incorporates a second set of ratio arms, often denoted as 'p' and 'q', in addition to the main ratio arms 'R1' and 'R2'. These auxiliary ratio arms are connected such that the resistance of the lead connecting the unknown resistor to the standard resistor (often denoted as 'R_L' or 'R_c') is effectively 'bridged out'. The galvanometer is connected between the appropriate points in the circuit so that the voltage drop across this connecting lead resistance does not appear across the galvanometer terminals at balance.

Specifically, the lead resistance connecting the unknown resistor to the standard resistor is placed within the bridge structure in such a way that its effect cancels out when the bridge is balanced, provided the ratios of the main arms and the auxiliary arms are equal. This clever arrangement ensures that the bridge measures only the true resistance of the unknown resistor, largely immune to the resistances of the leads and contacts.

Kelvin's Double Bridge vs. Wheatstone Bridge

Understanding the distinction between these two bridge types helps clarify why Kelvin's Double Bridge is superior for low resistance measurements:

Feature Wheatstone Bridge Kelvin's Double Bridge
Resistance Range Medium resistances (typically 1 Ohm to 1 MOhm) Very low resistances (typically < 1 Ohm)
Lead & Contact Errors Significant for low resistance measurements, leading to inaccuracies. Effectively eliminated due to its design, ensuring high accuracy.
Terminal Connections Uses two-terminal resistors. Uses four-terminal resistors (current and potential).
Complexity Simpler design. More complex design with additional ratio arms.

Conclusion on Error Elimination

Therefore, it is the Kelvin's Double bridge that specifically eliminates errors due to contacts and lead resistances in bridge measuring instruments, making it the preferred choice for highly accurate measurements of very low resistance values.

Was this answer helpful?

Important Questions from Measurement of Resistance

  1. Which of these instruments is used for measuring the insulation resistance?

  2. A ______ is a combined unit of ampere meter, voltmeter and ohmmeter.

  3. Which instrument is used to measure the high resistance?

  4. Megger is essentially a:

  5. Which one of the following does not employ a null method of measurement ?

Need Expert Advice?

Start Your Preparation with Prepp Mobile App

Download the app from Google Play & App Store
Download the app from Google Play & App Store
Prepp Mobile App