An instrument with an internal resistance of 100 Ω and a full-scale current of 1 mA is to be converted into a DC voltmeter with range of 0 V - 500 V. Find the value of the resistance used as a multiplier.  

The correct answer is 499900 Ω

Converting a Galvanometer to a DC Voltmeter

A galvanometer is a sensitive instrument that can detect small electric currents. To convert it into a DC voltmeter, a high resistance, known as a multiplier resistance, is connected in series with the galvanometer. This series resistance increases the overall resistance of the instrument and limits the current flowing through the galvanometer coil, allowing it to measure larger voltages.

The full-scale deflection current (\(I_g\)) of the galvanometer represents the maximum current the galvanometer can safely handle. When used as a voltmeter, this current will flow through the galvanometer and the series resistance when the maximum voltage (\(V\)) of the desired range is applied across the combination.

The total resistance of the voltmeter (\(R_v\)) is the sum of the galvanometer's internal resistance (\(R_g\)) and the series multiplier resistance (\(R_s\)):

\(R_v = R_g + R_s\)

According to Ohm's Law, the voltage across the voltmeter at full-scale deflection is given by:

\(V = I_g \times R_v\)

Substituting the expression for \(R_v\):

\(V = I_g (R_g + R_s)\)

We want to find the value of the multiplier resistance, \(R_s\). We can rearrange the formula to solve for \(R_s\):

\(\frac{V}{I_g} = R_g + R_s\)

\(R_s = \frac{V}{I_g} - R_g\)

Calculating the Multiplier Resistance

We are given the following values:

  • Internal resistance of the galvanometer, \(R_g = 100 \, \Omega\)
  • Full-scale current of the galvanometer, \(I_g = 1 \, \text{mA} = 1 \times 10^{-3} \, \text{A}\)
  • Desired full-scale voltage range, \(V = 500 \, \text{V}\)

Now, let's substitute these values into the formula for \(R_s\):

\(R_s = \frac{500 \, \text{V}}{1 \times 10^{-3} \, \text{A}} - 100 \, \Omega\)

\(R_s = 500 \times 10^3 \, \Omega - 100 \, \Omega\)

\(R_s = 500000 \, \Omega - 100 \, \Omega\)

\(R_s = 499900 \, \Omega\)

Thus, a series resistance of \(499900 \, \Omega\) is required to convert the given galvanometer into a DC voltmeter with a range of 0 V - 500 V.

Was this answer helpful?

Important Questions from Extension Ranges of Basic Meters

  1. A 1 mA, 100 Ω meter movement is converted to a 100 mA ammeter using a parallel shunt resistor. If this shunt resistor suddenly becomes open-circuited while a steady line current of 80 mA is passing through the meter, what is the most likely outcome?

  2. What is the value of series resistance required to extend the 0 - 100 Volts range of a 20,000 Ω/V meter to 0 - 1000 volts?

  3. A 100 micro Amp, 3000-ohm meter movement, the shunt resistance for double the current range is ____

  4. A galvanometer has an internal resistance of 100 ohm and a shunt resistance of 20 ohm. Find the mul tiplying factor.
  5. An analog voltmeter uses external multiplier settings. With a multiplier setting of 20 k$\Omega$, it reads 440 V and with a multiplier setting of 80 k$\Omega$, it reads 352 V. For a multiplier setting of 40 k$\Omega$, the voltmeter reads
Need Expert Advice?

Start Your Preparation with Prepp Mobile App

Download the app from Google Play & App Store
Download the app from Google Play & App Store
Prepp Mobile App