A piezoelectric transducer is used to measure force. The size of the piezoelectric crystal used in it is 6 mm × 6 mm × 2 mm, with a voltage sensitivity of 0.055 Vm/N. If the voltage developed is 120 V, find the force.
39.24 N
A piezoelectric transducer converts mechanical stress or force into an electrical charge or voltage. This property is utilized to measure applied forces. The amount of voltage generated is directly proportional to the applied force, mediated by the transducer's material properties and geometry.
In this problem, we are given the following information:
We need to find the force ($F$) applied to the transducer.
The voltage ($V$) developed across a piezoelectric crystal due to an applied force ($F$) is related by the following formula, considering the piezoelectric voltage coefficient ($g$), the area ($A$) over which the force is applied, and the thickness ($t$) of the crystal:
The relationship is given by:
$$ V = g \cdot \frac{F}{A} \cdot t $$
Where:
First, let's list the given values and convert them into standard SI units (meters, Newtons, Volts):
Now, calculate the area ($A$) and thickness ($t$) in square meters and meters, respectively:
Next, rearrange the formula $ V = g \cdot \frac{F}{A} \cdot t $ to solve for the force ($F$):
$$ F = \frac{V \cdot A}{g \cdot t} $$
Substitute the values into the rearranged formula:
$$ F = \frac{120 \text{ V} \times (36 \times 10^{-6} \text{ m}^2)}{(0.055 \text{ Vm/N}) \times (2 \times 10^{-3} \text{ m})} $$
Perform the calculation:
$$ F = \frac{120 \times 36 \times 10^{-6}}{0.055 \times 2 \times 10^{-3}} \text{ N} $$
$$ F = \frac{4320 \times 10^{-6}}{0.11 \times 10^{-3}} \text{ N} $$
$$ F = \frac{4320}{0.11 \times 10^{(-3 - (-6))}} \text{ N} $$
$$ F = \frac{4320}{0.11 \times 10^3} \text{ N} $$
$$ F = \frac{4320}{110} \text{ N} $$
$$ F = \frac{432}{11} \text{ N} $$
$$ F \approx 39.2727... \text{ N} $$
The calculated force is approximately 39.27 N. Comparing this result with the given options, we find that option 3 (39.24 N) is the closest value. The slight difference is likely due to rounding in the provided sensitivity value or the options themselves.
Which of the following can act as an inverse transducer?
Which of the following converts a physical parameter to an electrical signal?
For transducers, following statements are given:
(A) An optical interferometer is useful for measuring extremely small motions
(B) The damping ratio of a seismic instrument should be low for good dynamic performance
(C) A rate gyro is a relative motion measuring device
(D) A pneumatic motion transducer is non-linear over a wide range of motion
(E) A piezoelectric accelerometer can only be used for static motion measurement
Choose the correct answer from the options given below:
Choose the INCORRECT statement with respect to the use of electrical transducers.
Consider the following transducers:
1. LVDT
2. Piezoelectric
3. Thermocouple
4. Photovoltaic cell
5. Strain gauge
Which of these are active transducers?