Question

A closed Lissajous pattern observed on an oscilloscope in the $\mathbf{X} - \mathbf{Y}$ mode has three horizontal tangencies and two vertical tangencies in steady state. If the signal frequency on channel-$\mathbf{X}$ is $600\text{ Hz}$, then the signal-frequency on channel-$\mathbf{Y}$ is ___________ Hz.

The correct answer is
900

Lissajous Pattern Frequency Calculation

The ratio of the frequencies of the signals applied to the X and Y channels of an oscilloscope, when displayed as a Lissajous pattern, can be determined from the number of horizontal and vertical tangencies (points where the pattern touches the boundary of its envelope) observed in the steady-state pattern.

The relationship is given by the formula:

$ \frac{f_Y}{f_X} = \frac{\text{Number of horizontal tangencies}}{\text{Number of vertical tangencies}} $

Where:

  • $f_Y$ is the frequency of the signal on the Y-channel.
  • $f_X$ is the frequency of the signal on the X-channel.

Given Information

From the problem description:

  • Number of horizontal tangencies ($N_h$) = 3
  • Number of vertical tangencies ($N_v$) = 2
  • Signal frequency on channel-X ($f_X$) = $600\text{ Hz}$

Calculating Channel-Y Frequency

Substitute the known values into the formula:

$ \frac{f_Y}{600\text{ Hz}} = \frac{3}{2} $

To find $f_Y$, multiply both sides by $600\text{ Hz}$:

$ f_Y = 600\text{ Hz} \times \frac{3}{2} $

Perform the calculation:

$ f_Y = 300 \times 3 \text{ Hz} $

$ f_Y = 900\text{ Hz} $

Conclusion

The signal frequency on channel-Y is $900\text{ Hz}$.

Was this answer helpful?

Important Questions from Cathode Ray Oscilloscope

  1. In the following statements:
    A. Dual trace oscilloscope can be used in alternate and chop modes but not in x-y mode.
    B. Lissajous patterns are obtained in x-y mode
    C. Phosphor storage CRT have longer CRT life than mesh storage CRT.
    D. DC coupling is useful for low-frequency and low repetition rate signals.
    E. Sampling is done in cathode ray oscilloscopes.
    Choose the correct answer from the options given below:

  2. The shape of the time base signal of a CRO is generally a _____.
  3. A CRO displays a sine waveform which spans 3 cycles in 6 horizontal divisions. If the time base switch is set to 1 ms/div, what is the frequency of this signal which is being measured?

  4. The lowest scale probe suitable for measuring 250V DC using a CRO with I/P limit of 30V is:
  5. How is a Lissajous pattern created?